Datex-Ohmeda Aespire Troubleshooting
Listed below are some of the most common problems users run into using this device.
Drive Pressure Limit Switch Fails
Go to the Discrete 1/0 signal menu and check the DPL (Drive Pressure Limit) switch status (Off). Remove the rear subfloor. Connect a pressure test device (included with device setup) to the black inline connector that is in the Exhalation Valve interface cuff tubing (S18 side). Apply 104 +5/ -4 cm H20 (76.5 +3.8/-2.9 mm Hg) to activate the switch. Verify status change on the 1/0 signal screen (On). If problem continues, you will need to replace the S18.
Possible Leaking
If either sensor reads more than 2 LPM, STOP. This indicates a possible leak in the flow sensor pneumatic circuit. If the pressure is increased further, the ESIB (enhanced sensor interface board) may be damaged.
If the flow measurements on the ADC page stay near zero, the flow sensors are good. If either sensor indicates a flow where there is none, there may be a leak. Try reducing the circuit back to zero. Reverse the flow sensor connections at the ESIB interface panel and repeat test.
If the problem follows the sensor, discard the sensor. It has a leak. If the problem stays with the same side of the circuit, it is likely the leak is in the tubing, not with the sensor.
If the bleed resistor calibration test fails, check altitude and drive gas settings. Ensure that the breathing circuit gas is exhausting out the scavenging port during the test. A negative finding indicates a massive circuit leak. Check for proper regulator pressure calibration. Ensure that test was conducted properly. Otherwise, failure indicates a calibration problem with the flow valve.
High Pressure Leak
Listed below are several possible causes for a high pressure leak. If all tests are attempted and there is still a leak, contact technical support.
Pipeline Leak
Use a leak detector or Snoop (both included in device assembly) to check for source of leak. Repair or replace defective parts.
O2 Flush Valve
Use a leak detector or Snoop to check for source of leak. Make sure tubing connections are tight. Replace valve if defective.
System Switch
Use a leak detector or Snoop to check for source of leak. Make sure tubing connections are tight. Replace switch if defective.
Cylinder Not Installed Properly
Make sure cylinder is correctly aligned. verify that tee handles tight.
Cylinder Gauges
Use a leak detector or Snoop to check for source of leak. Replace gauge if defective.
Cylinder Gaskets
Use a leak detector or Snoop to check for source of leak. Replace gasket if defective.
Relief Valves
Use a leak detector or Snoop to check for source of leak. Replace valve if defective.
Low Pressure Leak
Listed below are several possible causes for a low pressure leak. If all tests are attempted and there is still a leak, contact technical support.
Vaporizer Not Installed Properly
Reset vaporizer if not installed properly. Have vaporizer serviced at vaporizer center if vaporizer leaks.
Missing or Damaged O-ring on Vaporizer Manifold
Check condition of O-ring. Replace if missing or damaged.
Loose Fill Port
Check fill port. Tighten if loose.
Breathing System Leak
Absorber Cannister Not Installed Properly
Remove cannister. Carefully adjust until cannister is locked in place.
Soda Lime Dust on Cannister Seals
Carefully dust seals and locking surfaces using a dry rag.