跳转到主内容
Philips Respironics V60 Battery Replacement
医疗设备
ifixit 或本维修指南均未获得医疗设备制造商的认可或与任何制造商存在附属关系。
你所需要的
  1. Philips Respironics V60 Battery Replacement, shut down the ventilator: 步骤 1 中的图像 1,2 Philips Respironics V60 Battery Replacement, shut down the ventilator: 步骤 1 中的图像 2,2
    • Press and release the ON/Shutdown key, opening the Shutdown window in the screen.
    • Select Ventilator Shutdown, which will shut down the ventilator.
    • Improper shutdown may cause a Power has been restored message the next time the ventilator is turned on.
    • Disconnect the ventilator from the AC power.
  2. Philips Respironics V60 Battery Replacement, remove side panel: 步骤 3 中的图像 1,1
    • The battery is located behind the left side panel of the device.
    • Remove the side panel by turning the captive Phillips head fastener a 1⁄4 turn and releasing.
  3. Philips Respironics V60 Battery Replacement, loosen battery bracket: 步骤 4 中的图像 1,1
    • Using a 3-mm hex wrench, remove the battery bracket by removing two screws.
  4. Philips Respironics V60 Battery Replacement, replace battery: 步骤 5 中的图像 1,1
    • Disconnect the battery cable by pinching the battery connector and pulling it out.
    • Remove the battery and bracket.
    • Holding the battery so that the vent hole faces up and the Philips logo faces out, thread the battery cable through the battery bracket. Position and place the battery inside the battery compartment.
    • Pinching the end of the battery connector, plug it in so that it locks in place.
    • Reinstall the battery bracket by fastening the two screws with the 3mm hex key.
    • Reinstall the side panel and secure the fastener with a 1⁄4 turn clockwise.
  5. Philips Respironics V60 Battery Replacement, charging: 步骤 7 中的图像 1,1
    • Connect the ventilator back to the AC power.
    • Make sure the battery is properly installed by verifying that the yellow Battery (charged) LED on the front panel flashes. The flashing LED indicates the battery is being charged.
    • A new battery must be charged for at least 5 hours before being placed into service.
    • The manufacturer recommends that the ventilator’s batteries be fully charged before ventilating a patient. If the batteries are not fully charged and AC power fails, always pay close attention to the level of battery charge.
    • The ventilator charges the battery whenever the ventilator is connected to AC, with or without the ventilator switched on. The Battery (charged) LED flashes to show that the battery is being charged.
    • Power on the ventilator with the ON/Shutdown key.
    • Run the preoperational check.
    • Following battery installation, if a Check Vent or Vent Inoperative alarm occurs during the preoperational check, discontinue use of the ventilator immediately and contact Respironics. The Vent Inoperative alarm occurs if AC power is disconnected and a battery is not installed, or if the battery is fully discharged.
终点
knipwim
于09/01/18注册
8,527 声望
创作了30篇指南
0条评论
添加评论
浏览统计数据:
过去 24 小时: 1
过去 7 天: 5
过去 30 天: 32
总计 2,358
