跳转到主内容
Philips Emerson CoughAssist CA Series Testing Procedure
医疗设备
ifixit 或本维修指南均未获得医疗设备制造商的认可或与任何制造商存在附属关系。
  1. Philips Emerson CoughAssist CA Series Testing Procedure, Philips Emerson CoughAssist CA Series Testing Procedure: 步骤 1 中的图像 1,1
    • For CM models, the pressure knob is located on the left in place of the timing knobs pictured here.
    • Procedure can be found in the attached document.
    • Block the connector on the front of the machine using a rubber plug or similar item.
    • Check the pressure gauge. If it is not zero, see “pressure gauge reading is wrong” on the troubleshooting guide.
    • Set the inhale pressure knob fully clockwise, and the inhale flow and pressure knobs fully counterclockwise. On the CA models, set the manual/auto switch to manual, and set each timing knob to 1 second.
    • Turn on the power switch, checking that the light under it turns on. If not, see “indicator light won’t turn on” in the troubleshooting guide.
    • Turn the pressure knob slowly clockwise, noting that the speed of the motor increases steadily. Once it’s fully clockwise, allow the unit to warm up for 1 to 2 minutes.
    • Reset the pressure knob counterclockwise, then set the control lever to inhale and check that the gauge reads between 2 and 5 cmH2O.
    • Move the control lever to exhale and note that the gauge reads between -2 and -5 cmH2O.
    • Repeat the inhale and exhale tests with the pressure knob fully clockwise, seeing if the gauge is between 57 and 63 cmH2O.
    • Release the control lever and note that the lever moves to the center and pressure drops immediately to 0 cmH2O.
    • If any of these pressures are wrong, see “limited pressure” on the troubleshooting guide.
    • With the control lever on inhale, adjust the pressure knob to 40 cmH2O.
    • Turn the inhale pressure knob fully counterclockwise. Check that the gauge is between 18 and 22 cmH2O.
    • Slowly returning the inhale pressure knob clockwise should increase the pressure gradually back to 40 cmH2O.
    • 5. Turn the pressure knob fully clockwise, set the control lever to inhale, remember the gauge reading, and release the control lever.
    • Set inhale flow to minimum (clockwise). Set the control lever to inhale and check that the pressure is within 10 cmH2O below the previous reading (maximum flow).
    • Remove the rubber plug. Check the flow by pushing the control lever to inhale and feel the airflow out of the unit. Release the control lever.
    • Set the inhale flow knob fully counterclockwise, and repeat the test, checking for a large increase of airflow.
    • (The rest of this guide only applies to the CA series, not the CM series.)
    • Set the pressure knob to minimum. Move the switch from manual to automatic.
    • Verify that the unit immediately begins with the inhale pressure, followed by exhale (negative) pressure then drops to 0 cmH2O and continues to repeat.
    • Set the switch back to manual and verify that the pressure drops to 0 cmH2O. Return to auto.
    • One at a time, set each timing knob to 5 seconds and verify that each knob affects the correct phase.
    • Set the timing knobs to 5 seconds and use a stopwatch to check that each one lasts between 4.8 and 5.2 seconds.
    • Set the knobs to 1 second and verify that each phase lasts between 0.9 and 1.1 seconds.
    • If the timings are off, see “incorrect timings” on the troubleshooting guide.
终点
团队
Cal Poly, Team S11-G18, Paton Spring 2020 Cal Poly, Team S11-G18, Paton Spring 2020 的会员
CPSU-PATON-S20S11G18
3 名成员
创作了3篇指南
0条评论
添加评论
浏览统计数据:
过去 24 小时: 1
过去 7 天: 1
过去 30 天: 20
总计 307
