    • Set the gas supply to 5 L/min.
  1. How to Test the Valve System on a Fisher Paykel Neopuff: 步骤 2 中的图像 1,1
    • Completely close the Peak Inspiratory Pressure control and the Max Pressure control by turning both knobs completely clockwise.
  2. How to Test the Valve System on a Fisher Paykel Neopuff: 步骤 3 中的图像 1,1
    • Close the bleed valve and check that the gauge reads at least 60 cmh2O.
    • Set the gas supply to 15 L/min.
  3. How to Test the Valve System on a Fisher Paykel Neopuff: 步骤 5 中的图像 1,1
    • Close the bleed valve and check that the gauge reads no higher than 80 cmH2O.
    • Set the gas supply to 10 L/min.
    • Close the bleed valve and turn the Max Pressure Relief valve until the manometer reads 40 cmh2O.
    • Check that the manometer needle rises and falls smoothly.
    • Gently rotate the Max Pressure Relief cover over the Max Pressure Relief knob.
  4. How to Test the Valve System on a Fisher Paykel Neopuff: 步骤 10 中的图像 1,1
    • Reset the peak inspiratory pressure to 20 cmh2O and turn off the gas flow. Testing is now complete. If the F&P Neopuff/Perivent Infant Resuscitator fails any of these tests, the valve assembly should be regarded as faulty and replaced with a new valve assembly (see the parts list (section 4.1.2) for part numbers).