SEPTEMBE GOMCO® MOBILE CONSTANT AND INTERMITTENT MODELS 6036 & 6037

OPERATION, MAINTENANCE AND SERVICE MANUAL

INDEX

1.0 General Information

2.0 Specifications

3.0 Set-Up

4.0 Operating Procedure

5.0 Operating Principle

6.0 Maintenance and Service

7.0 Illustrations

8.0 Trouble Shooting

9.0 Replacement Parts List

10.0 Warranty

11.0 Copyright

1. 0 GENERAL INFORMATION:

1.1 Purpose:

The Gomco Models 6036 and 6037 Constant and Intermittent pumps are designed to provide two types of vacuum service. Constant suction provides vacuum levels fully adjustable from 0 to 200 mm of mercury with a maximum air flow of 25 liters per minute. For applications that require mild intermittent suction, the pump will provide 0 to 200 mm of mercury with a maximum air flow of 4 liters per minute. While using intermittent suction, the user can adjust the amount of time the vacuum is to be applied (1 to 90 seconds). The amount of time the unit vents to atmosphere is 12 seconds.

1.2 Features, Model 6036:

Supplied with Model 6036 as standard equipment are:

  1. One (1) 600 ml glass collection bottle (used on constant).

  2. One (1) 2800 ml glass collection bottle with overflow protection cap assembly (used on intermittent).

  3. Two (2) disposable tubing packages, each containing a 15-inch and a 72-inch length of clean 1/4 inch I.D. clear flexible PVC tubing.

  4. Two (2) packages, of 3 each, of bacteria filters.

  5. Compact cart with non-marking casters, easy to clean stainless steel top work surface, illuminated on-off switch, lamps to indicate constant or intermittent service, "on" time control, vacuum gauge, vacuum regulator and hospital grade plug.

1.2.1 2800 ml Bottle Assembly:

Your Gomco Pump has an overflow protection device called the Cap and Float Assembly. When collected patient drainage fluids reach the last graduation of your collection bottle, the vacuum supply to the collection bottle is shut off by the cap and float assembly.

1.3 Features, Model 6037

Supplied with Model 6037 as standard equipment are:

  1. One (1) 1100 ml disposable collection container with overflow protection cap (used on constant).

  2. One (1) 2100 ml disposable collection container with overflow protection cap (used on intermittent).

  3. Two (2) disposable tubing packages, each containing a 15-inch and a 72-inch length of clean 1/4 inch I.D. clear flexible PVC tubing.

  4. Two (2) packages, of 3 each, of the bacterial filters.

  5. Compact cart with non-marking casters, easy to clean stainless steel top work surface, illuminated on-off switch, lamps to indicate constant or intermittent service, "on" time control, vacuum gauge, vacuum regulator and hospital grade plug.

1.4 Bacterial Filter:

The high efficiency bacterial filter is common to both the 6036 and 6037 units. The filter is custom engineered to prevent fluid and aerosol contamination of mobile suction units. This filter features a hydrophobic, micro-porous membrane which filters air with maximum efficiency (0.3 micron particles in air), while blocking the flow of aqueous fluids and aerosol contaminants. The Gomco high efficiency filter protects against suction pumps contamination in the case of canister overflow as it helps prevent the overflow from reaching the pump.

2. 0 SPECIFICATIONS:

NOTE: Specifications are nominal and subject to change without prior notice.

2.1 VACUUM RANGE: 0 to 200 mm Hg.

2.2 FLOW RATE (Open): Constant - 20 lpm Intermittent - 4 lpm

2.3 ELECTRICAL REQUIREMENTS:

115 volts 60 Hz (2.0 amps)

(A 230 volt unit is also available)

2.4 MOTOR AND PUMP DESCRIPTION:

4 pole, permanent split capacitor motor, direct drive diaphragm pump.

2.4.1 PUMP (Only):

Vacuum - 22 inch Hg. Minimum

Flow - 28 lpm

2.5 DIMENSIONS:

(H) 33 in. X (D) 17 in. X (W) 23 in.

2.6 WEIGHT:

Shipping - 78 Pounds

Net - 62 Pounds

2.7 DUTY:

Continuous: 24 hours

3. 0 SET-UP:

3.1 BACTERIA FILTERS:

Install two bacteria filters to the stand prior to patient applications as follows.

  1. Locate the vacuum and filter connection on the bottom of the stand body, (See figure 1).

  2. Screw the threaded ends of the filter into the vacuum filter connections by turning in a clockwise direction.

  3. Tighten finger tight.

CAUTION: Care must be taken when installing the filter to prevent thread damage due to cross threading and over tightening.

Image

BACTERIA FILTER tion Bottle

3.2 600 ml Collection Bottle (See Illustration No. 5)

  1. The bottle and cap assembly should be sterilized prior to use (See 6.4).

  2. The bottle should be assembled as shown in illustration No. 5.

  3. Insert the bottle assembly into the wire bottle bracket located on the base of the stand.

  4. Connect the 15-inch length of tubing, enclosed in the tubing package, to the "short" metal bottle tube and to the bacteria filter located under the word "CONSTANT".

  5. Connect the 72-inch length of patient tubing, enclosed in the tubing package, to the "long" metal bottle tube.

3.3 2800 ml Collection Bottle:

(See Illustration No. 6)

  1. The bottle and cap assembly should be sterilized prior to use (see 6.4).

  2. The bottle should be assembled as shown in Illustration No. 6.

  3. Insert the bottle assembly into the wire bottle bracket located on the base of the stand.

  4. Connect the 15-inch length of tubing, enclosed in the tubing package, to the vertical fitting on the bottle cap and float assembly marked "to pump" and to the bacteria filter located under the word "INTERMITTENT".

  5. Connect the 72-inch length of patient tubing to the long bent metal tube of the cap assembly.

3.4 1100 ml Disposable Collection Container -